Basic SEROUT SERIN question
Archiver
Posts: 46,084
I have what i think is a very basic question. I am trying to use a
Basic stamp 2 to controll a water bath/Circulator(PolyScience 9112).
It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
send out a comand to turn on the bath i.e.:
SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
That works fine, but the problem lies on the SERIN. When a camand
such as the one above is sent the bath replys with an ASCII '!' to
say that its been recived. so I'm trying to do the following:
SerData VAR BYTE
SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
PAUSE 25
SERIN 1, 16780,[noparse][[/noparse]SerData]
DEBUG SerData
OR
SerData VAR BYTE
SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
SERIN 1, 16780,[noparse][[/noparse]SerData]
DEBUG SerData
The problem is i never see the ! in the debug window or even when i
have hyperterminal connected in parallel.
Idealy I want to be able to send out a that will respond in the
temprature 'xxx.xx'
SO anyone out there have any idea why I'm not seeing that '!' comming
back?
Basic stamp 2 to controll a water bath/Circulator(PolyScience 9112).
It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
send out a comand to turn on the bath i.e.:
SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
That works fine, but the problem lies on the SERIN. When a camand
such as the one above is sent the bath replys with an ASCII '!' to
say that its been recived. so I'm trying to do the following:
SerData VAR BYTE
SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
PAUSE 25
SERIN 1, 16780,[noparse][[/noparse]SerData]
DEBUG SerData
OR
SerData VAR BYTE
SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
SERIN 1, 16780,[noparse][[/noparse]SerData]
DEBUG SerData
The problem is i never see the ! in the debug window or even when i
have hyperterminal connected in parallel.
Idealy I want to be able to send out a that will respond in the
temprature 'xxx.xx'
SO anyone out there have any idea why I'm not seeing that '!' comming
back?
Comments
Rpin is a variable/constant/expression* (0 - 16) that specifies the
I/O pin through which the serial data will be received. This pin will
be set to input mode. On the BS2, BS2e, BS2sx, BS2p and BS2pe, if
Rpin is set to 16, the BASIC Stamp uses the dedicated serial-input
pin (SIN, physical pin 2), which is normally used by the Stamp Editor
during the download process.
Did you mean to use an Rpin value of 16 instead of 1?
If you are using the dedicated serial i/o use 16. Otherwise, I
believe you need to be using two different pins.
I am new to the Stamp, but it just caught me off guard that you were
using the same pin for input and output.
Good luck,
Frank
--- In basicstamps@yahoogroups.com, "pueet" <pueet@y...> wrote:
> I have what i think is a very basic question. I am trying to use a
> Basic stamp 2 to controll a water bath/Circulator(PolyScience
9112).
> It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
> send out a comand to turn on the bath i.e.:
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> That works fine, but the problem lies on the SERIN. When a camand
> such as the one above is sent the bath replys with an ASCII '!' to
> say that its been recived. so I'm trying to do the following:
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> PAUSE 25
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> OR
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> The problem is i never see the ! in the debug window or even when i
> have hyperterminal connected in parallel.
>
> Idealy I want to be able to send out a that will respond in the
> temprature 'xxx.xx'
>
> SO anyone out there have any idea why I'm not seeing that '!'
comming
> back?
You don't want the pause between the SERIN and SEROUT commands. You said
that the unit you are connecting to can use ether inverted or true. Do you
know which it returns? Is it the same as is transmitted to it, or does it
have a jumper to set it, or is it always one or the other? So you might try
396 as the baud rate on the SERIN and see if it is sending back true data.
Another guess might be that the ACK (the "!") is too fast for the Stamp to
pick up. Do you have access to an oscilliscope? That way you could see if
the ACK is being sent or not.
Jonathan
www.madlabs.info
Original Message
From: "pueet" <pueet@y...>
To: <basicstamps@yahoogroups.com>
Sent: Thursday, June 10, 2004 2:14 PM
Subject: [noparse][[/noparse]basicstamps] Basic SEROUT SERIN question
> I have what i think is a very basic question. I am trying to use a
> Basic stamp 2 to controll a water bath/Circulator(PolyScience 9112).
> It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
> send out a comand to turn on the bath i.e.:
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> That works fine, but the problem lies on the SERIN. When a camand
> such as the one above is sent the bath replys with an ASCII '!' to
> say that its been recived. so I'm trying to do the following:
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> PAUSE 25
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> OR
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> The problem is i never see the ! in the debug window or even when i
> have hyperterminal connected in parallel.
>
> Idealy I want to be able to send out a that will respond in the
> temprature 'xxx.xx'
>
> SO anyone out there have any idea why I'm not seeing that '!' comming
> back?
>
>
>
>
> To UNSUBSCRIBE, just send mail to:
> basicstamps-unsubscribe@yahoogroups.com
> from the same email address that you subscribed. Text in the Subject and
Body of the message will be ignored.
>
> Yahoo! Groups Links
>
>
>
>
>
>
You want to say:
SERIN 1,16780, [noparse][[/noparse]STR SerData\1]
Which means, get from pin 1 a single byte, as a
'string'.
When you get '!' with a 'plain' SERIN, SERIN
recieves that as a 0..255 value -- actually,
it recieves the ascii value of '!' which is
33. So, I assume your 'debug' output is 33,
which is the return character.
Now, you also must know that IF the response
begins to be sent BEFORE you issue your
'SERIN', then the first few bits may be
missed by the BS2. So you want to issue your
SERIN as soon as possible -- and even that
may not be soon enough.
Note the SERIN will 'pend', waiting for a
byte to be recieved, forever if you don't
put a time-out on it.
--- In basicstamps@yahoogroups.com, "pueet" <pueet@y...> wrote:
> I have what i think is a very basic question. I am trying to use a
> Basic stamp 2 to controll a water bath/Circulator(PolyScience
9112).
> It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
> send out a comand to turn on the bath i.e.:
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> That works fine, but the problem lies on the SERIN. When a camand
> such as the one above is sent the bath replys with an ASCII '!' to
> say that its been recived. so I'm trying to do the following:
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> PAUSE 25
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> OR
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> The problem is i never see the ! in the debug window or even when i
> have hyperterminal connected in parallel.
>
> Idealy I want to be able to send out a that will respond in the
> temprature 'xxx.xx'
>
> SO anyone out there have any idea why I'm not seeing that '!'
comming
> back?
Thanks for the reply, just as you suggested I connected a
oscilliscope to the TX and RX lines and send out the command. just
as expected i recived a signal back at about 5ms after the last data
signal on the transmit. so once again i'm stuck at why don't i see
the information back. even tried the 396 as you suggested. one
thing i did noticed is that the signal i send is at a peak of about
4.3V where the signal coming to the stamp is at about 6v. could that
be a factor? any other suggestions?
Justin
--- In basicstamps@yahoogroups.com, "Jonathan Peakall"
<jpeakall@p...> wrote:
> Pueet,
>
> You don't want the pause between the SERIN and SEROUT commands. You
said
> that the unit you are connecting to can use ether inverted or true.
Do you
> know which it returns? Is it the same as is transmitted to it, or
does it
> have a jumper to set it, or is it always one or the other? So you
might try
> 396 as the baud rate on the SERIN and see if it is sending back
true data.
>
> Another guess might be that the ACK (the "!") is too fast for the
Stamp to
> pick up. Do you have access to an oscilliscope? That way you could
see if
> the ACK is being sent or not.
>
> Jonathan
>
> www.madlabs.info
>
>
>
Original Message
> From: "pueet" <pueet@y...>
> To: <basicstamps@yahoogroups.com>
> Sent: Thursday, June 10, 2004 2:14 PM
> Subject: [noparse][[/noparse]basicstamps] Basic SEROUT SERIN question
>
>
> > I have what i think is a very basic question. I am trying to use
a
> > Basic stamp 2 to controll a water bath/Circulator(PolyScience
9112).
> > It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able
to
> > send out a comand to turn on the bath i.e.:
> > SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> > That works fine, but the problem lies on the SERIN. When a camand
> > such as the one above is sent the bath replys with an ASCII '!' to
> > say that its been recived. so I'm trying to do the following:
> >
> > SerData VAR BYTE
> >
> > SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> > PAUSE 25
> > SERIN 1, 16780,[noparse][[/noparse]SerData]
> > DEBUG SerData
> >
> > OR
> >
> > SerData VAR BYTE
> >
> > SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> > SERIN 1, 16780,[noparse][[/noparse]SerData]
> > DEBUG SerData
> >
> > The problem is i never see the ! in the debug window or even when
i
> > have hyperterminal connected in parallel.
> >
> > Idealy I want to be able to send out a that will respond in the
> > temprature 'xxx.xx'
> >
> > SO anyone out there have any idea why I'm not seeing that '!'
comming
> > back?
> >
> >
> >
> >
> > To UNSUBSCRIBE, just send mail to:
> > basicstamps-unsubscribe@yahoogroups.com
> > from the same email address that you subscribed. Text in the
Subject and
> Body of the message will be ignored.
> >
> > Yahoo! Groups Links
> >
> >
> >
> >
> >
> >
You say that you have checked the TX/RX pins, so it sounds like they are
seperate pins (one for TX, on for RX) but your code lookes at the same pin
for both TX and RX. Is this correct?
I would think the Stamp would tolerate the extra 1V, but I could be wrong.
Theroretically one shouldn't ever input more than 5V. If that were the case,
maybe trying a RS232 chip would make it work, as this would change the
voltage to 5V. Still I don't think that is it.
I assume that your code is hanging when waiting for the ACK? I would also
think that 5mS is enough time for the Stamp to get the ACK. Any experts out
there know if this is true?
Jonathan
www.madlabs.info
Original Message
From: "pueet" <pueet@y...>
To: <basicstamps@yahoogroups.com>
Sent: Friday, June 11, 2004 12:51 PM
Subject: [noparse][[/noparse]basicstamps] Re: Basic SEROUT SERIN question
> Jonathan,
> Thanks for the reply, just as you suggested I connected a
> oscilliscope to the TX and RX lines and send out the command. just
> as expected i recived a signal back at about 5ms after the last data
> signal on the transmit. so once again i'm stuck at why don't i see
> the information back. even tried the 396 as you suggested. one
> thing i did noticed is that the signal i send is at a peak of about
> 4.3V where the signal coming to the stamp is at about 6v. could that
> be a factor? any other suggestions?
>
> Justin
>
> --- In basicstamps@yahoogroups.com, "Jonathan Peakall"
> <jpeakall@p...> wrote:
> > Pueet,
> >
> > You don't want the pause between the SERIN and SEROUT commands. You
> said
> > that the unit you are connecting to can use ether inverted or true.
> Do you
> > know which it returns? Is it the same as is transmitted to it, or
> does it
> > have a jumper to set it, or is it always one or the other? So you
> might try
> > 396 as the baud rate on the SERIN and see if it is sending back
> true data.
> >
> > Another guess might be that the ACK (the "!") is too fast for the
> Stamp to
> > pick up. Do you have access to an oscilliscope? That way you could
> see if
> > the ACK is being sent or not.
> >
> > Jonathan
> >
> > www.madlabs.info
> >
> >
> >
Original Message
> > From: "pueet" <pueet@y...>
> > To: <basicstamps@yahoogroups.com>
> > Sent: Thursday, June 10, 2004 2:14 PM
> > Subject: [noparse][[/noparse]basicstamps] Basic SEROUT SERIN question
> >
> >
> > > I have what i think is a very basic question. I am trying to use
> a
> > > Basic stamp 2 to controll a water bath/Circulator(PolyScience
> 9112).
> > > It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able
> to
> > > send out a comand to turn on the bath i.e.:
> > > SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> > > That works fine, but the problem lies on the SERIN. When a camand
> > > such as the one above is sent the bath replys with an ASCII '!' to
> > > say that its been recived. so I'm trying to do the following:
> > >
> > > SerData VAR BYTE
> > >
> > > SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> > > PAUSE 25
> > > SERIN 1, 16780,[noparse][[/noparse]SerData]
> > > DEBUG SerData
> > >
> > > OR
> > >
> > > SerData VAR BYTE
> > >
> > > SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> > > SERIN 1, 16780,[noparse][[/noparse]SerData]
> > > DEBUG SerData
> > >
> > > The problem is i never see the ! in the debug window or even when
> i
> > > have hyperterminal connected in parallel.
> > >
> > > Idealy I want to be able to send out a that will respond in the
> > > temprature 'xxx.xx'
> > >
> > > SO anyone out there have any idea why I'm not seeing that '!'
> comming
> > > back?
> > >
> > >
> > >
> > >
> > > To UNSUBSCRIBE, just send mail to:
> > > basicstamps-unsubscribe@yahoogroups.com
> > > from the same email address that you subscribed. Text in the
> Subject and
> > Body of the message will be ignored.
> > >
> > > Yahoo! Groups Links
> > >
> > >
> > >
> > >
> > >
> > >
>
>
>
>
> To UNSUBSCRIBE, just send mail to:
> basicstamps-unsubscribe@yahoogroups.com
> from the same email address that you subscribed. Text in the Subject and
Body of the message will be ignored.
>
> Yahoo! Groups Links
>
>
>
>
>
>
> I have what i think is a very basic question. I am trying to use a
> Basic stamp 2 to controll a water bath/Circulator(PolyScience
9112).
> It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
> send out a comand to turn on the bath i.e.:
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> That works fine, but the problem lies on the SERIN. When a camand
> such as the one above is sent the bath replys with an ASCII '!' to
> say that its been recived. so I'm trying to do the following:
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> PAUSE 25
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> OR
>
> SerData VAR BYTE
>
> SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> SERIN 1, 16780,[noparse][[/noparse]SerData]
> DEBUG SerData
>
> The problem is i never see the ! in the debug window or even when i
> have hyperterminal connected in parallel.
>
> Idealy I want to be able to send out a that will respond in the
> temprature 'xxx.xx'
>
> SO anyone out there have any idea why I'm not seeing that '!'
comming
> back?
At the risk of adding more confusion to your life, may I suggest
the following:
1. You are using the same pin to both transmit and receive. Use a
different pin for each unless you use pin 16 ( as has been pointed
out ) If you do use pin 16, the baud parameter will be 396.
2. Don't put the pause between Serout and serin. Once you do the
serout, immediately setup the serin in the next line.
3. Most importantly, use the 'WAIT' parameter so serin will wait
around until it sees the !. What is happpening is that unless the !
shows up at the instant the serin line executes, you miss the
character.
I'd suggest serin 2,16780,2000,nochar,[noparse][[/noparse]wait ("!")]
Normally, you'd also receive other characters, but in your case,
the only thing expected is the !. "nochar" is the label of the point
in the program that executes if no characters are received in 2 secs.
4. If you want confirmation on the debug screen, just add a debug
statement after the serin line. debug "!"
Tom Sisk
switching gears just a little bit here. you say your turning on bath water
via the stamp. you must have some type of water soleniods. Don't mind me
asking but where did you get them? I have been looking everywhere trying to
find 12V 100psi water valves. found some someplace but they wanted an 2
arms and 1 leg. maybe I'm wrong but thinking you can pick'em up for around
$30.
If you or anyone else knows where I can locate them please let me know.
thanks
Daryl
At 04:14 PM 6/10/2004, you wrote:
>I have what i think is a very basic question. I am trying to use a
>Basic stamp 2 to controll a water bath/Circulator(PolyScience 9112).
>It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
>send out a comand to turn on the bath i.e.:
>SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
>That works fine, but the problem lies on the SERIN. When a camand
>such as the one above is sent the bath replys with an ASCII '!' to
>say that its been recived. so I'm trying to do the following:
>
>SerData VAR BYTE
>
>SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
>PAUSE 25
>SERIN 1, 16780,[noparse][[/noparse]SerData]
>DEBUG SerData
>
>OR
>
>SerData VAR BYTE
>
>SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
>SERIN 1, 16780,[noparse][[/noparse]SerData]
>DEBUG SerData
>
>The problem is i never see the ! in the debug window or even when i
>have hyperterminal connected in parallel.
>
>Idealy I want to be able to send out a that will respond in the
>temprature 'xxx.xx'
>
>SO anyone out there have any idea why I'm not seeing that '!' comming
>back?
>
>
>
>
>To UNSUBSCRIBE, just send mail to:
> basicstamps-unsubscribe@yahoogroups.com
>from the same email address that you subscribed. Text in the Subject and
>Body of the message will be ignored.
>
>Yahoo! Groups Links
>
>
>
>
As for your question about the soleniods, you may have miss
understood me when i say "water bath" what it is is this unit:
http://www.polyscience.com/lab/9100.html
It is used in chemical labs to put a test sample in to raise or lower
it to a X deg temp. to see its shelf life or preasure it may put
off. all it is, is a 6 liter tank that is filled with a liquid such
as Ethylene Glycol (Anti-freeze) and a motor sucks the liquid up and
pumps it back (Circulate). so there are no valves or soleniods in
the unit. but you may whant to check through some of there other
products and see if they have something that may to fit your need and
contact them.
Good Luck,
Justin
--- In basicstamps@yahoogroups.com, Daryl Berryhill <djberry@h...>
wrote:
> Justin
>
> switching gears just a little bit here. you say your turning on
bath water
> via the stamp. you must have some type of water soleniods. Don't
mind me
> asking but where did you get them? I have been looking everywhere
trying to
> find 12V 100psi water valves. found some someplace but they wanted
an 2
> arms and 1 leg. maybe I'm wrong but thinking you can pick'em up for
around
> $30.
> If you or anyone else knows where I can locate them please let me
know.
>
> thanks
> Daryl
>
>
> At 04:14 PM 6/10/2004, you wrote:
> >I have what i think is a very basic question. I am trying to use a
> >Basic stamp 2 to controll a water bath/Circulator(PolyScience
9112).
> >It uses Rs-232 (2400 8-N-1 Inverted or non-inverted) I am able to
> >send out a comand to turn on the bath i.e.:
> >SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> >That works fine, but the problem lies on the SERIN. When a camand
> >such as the one above is sent the bath replys with an ASCII '!' to
> >say that its been recived. so I'm trying to do the following:
> >
> >SerData VAR BYTE
> >
> >SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> >PAUSE 25
> >SERIN 1, 16780,[noparse][[/noparse]SerData]
> >DEBUG SerData
> >
> >OR
> >
> >SerData VAR BYTE
> >
> >SEROUT 1, 16780,[noparse][[/noparse]83,79,49,13]
> >SERIN 1, 16780,[noparse][[/noparse]SerData]
> >DEBUG SerData
> >
> >The problem is i never see the ! in the debug window or even when i
> >have hyperterminal connected in parallel.
> >
> >Idealy I want to be able to send out a that will respond in the
> >temprature 'xxx.xx'
> >
> >SO anyone out there have any idea why I'm not seeing that '!'
comming
> >back?
> >
> >
> >
> >