Hi, I would like to make an automated test equipment to be used for
measuring resistance of kelvin contactors.

there ar 15 contactors so I want to measure the resistance of each
contactors which is less than 1 Ohm. I will conduct a pin to pin
measurement and a pin to ground measurement to check if the pins are
grounded.

Any ideas if how i can make this possible? The resistance involved is
less than 1 Ohm.

Thanks you.

Comments

  • ArchiverArchiver Posts: 46,084
    edited 2003-08-10 16:12
    Hello,

    I built a similar piece of test equipment a few years
    back. It was an automated power supply tester, but was
    very complex and used a PC and VisualBasic instead of
    a Stamp.

    I used a GPIB equipped DMM for the measuring portion
    of the project. I then built a multiplexor system with
    high-quality relays to direct the desired measurement
    at the appropriate time to the DMM. The relays were
    driven by a digital I/O PCI card, and the GPIB
    interface was a PCI card too.

    The program would increment the relay, hold it, take a
    measurement with the DMM, store the value, increment
    to the next relay, and so on. Once the sequence was
    complete, the list of values was compared for accuracy
    and the results displayed to an operator who passed or
    failed the power supply.

    My program was designed to be quick and very automated
    so as not to require a skilled operator. The operator
    installed ten power supplies in bays one through ten,
    and pushed start. Three minutes later, the operator
    was instructed to pass or fail certain slots and
    continue.

    If you are looking for something relatively slow, you
    may be able to skip the GPIB and PCI section and
    simply have a Stamp increment a relay matrix,
    connected to a DMM, each time a button is pressed. The
    operator (or yourself) would push the button, read the
    DMM (one with low resistance capabilities, or course),
    and press the button again, read it again, and so on.

    Cool project.

    Regards,
    Dr. Diode
    --- catcat <semiconductor@p...> wrote:
    > Hi, I would like to make an automated test equipment
    > to be used for
    > measuring resistance of kelvin contactors.
    >
    > there ar 15 contactors so I want to measure the
    > resistance of each
    > contactors which is less than 1 Ohm. I will conduct
    > a pin to pin
    > measurement and a pin to ground measurement to check
    > if the pins are
    > grounded.
    >
    > Any ideas if how i can make this possible? The
    > resistance involved is
    > less than 1 Ohm.
    >
    > Thanks you.
    >
    >
    > To UNSUBSCRIBE, just send mail to:
    > basicstamps-unsubscribe@yahoogroups.com
    > from the same email address that you subscribed.
    > Text in the Subject and Body of the message will be
    > ignored.
    >
    >
    > Your use of Yahoo! Groups is subject to
    > http://docs.yahoo.com/info/terms/
    >
    >


    __________________________________
    Do you Yahoo!?
    Yahoo! SiteBuilder - Free, easy-to-use web site design software
    http://sitebuilder.yahoo.com
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-10 17:56
    Your right this is a cool project, the only problem I see is that you are
    using a relay to measure relay contact resistance of another relay, even
    though the selecting relay is high quality it doesn't sound good.

    Is there a non mechanical solution anyone can see.

    I'm working on a similar project and I worry about relay noise (clicking)
    and relay life if you wanted to do continues monitoring.

    <snips>
    >
    > The program would increment the relay, hold it, take a
    > measurement with the DMM, store the value, increment
    > to the next relay, and so on. Once the sequence was
    > complete, the list of values was compared for accuracy
    > and the results displayed to an operator who passed or
    > failed the power supply.

    > Regards,
    > Dr. Diode
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-10 18:19
    For very low-resistance measurement, a good way to go is a 4-wire Kelvin
    Bridge, which nulls out the resistance of the test leads themselves. That
    way you can accurately measure resistance to below 0.1 ohms without having
    to resort to jumper cables for test leads. It does this by comparing the
    test resistance to a switchable bank of know-value resistors, and when the
    needle in the meter nulls, the test resistance equals the comparison
    resistance. We used this method to test and select resistors down to a few
    ppm (parts per million) accuracy. If you need to switch banks of test
    subjects, just use standard 4-pole relays.

    Mike Sokol
    www.modernrecording.com
    mikes@m...


    Original Message
    From: "Richard Furniss" <rfurniss@e...>
    To: <basicstamps@yahoogroups.com>
    Sent: Sunday, August 10, 2003 12:56 PM
    Subject: Re: [noparse][[/noparse]basicstamps] Multiplexed resistance meter


    > Your right this is a cool project, the only problem I see is that you are
    > using a relay to measure relay contact resistance of another relay, even
    > though the selecting relay is high quality it doesn't sound good.
    >
    > Is there a non mechanical solution anyone can see.
    >
    > I'm working on a similar project and I worry about relay noise (clicking)
    > and relay life if you wanted to do continues monitoring.
    >
    > <snips>
    > >
    > > The program would increment the relay, hold it, take a
    > > measurement with the DMM, store the value, increment
    > > to the next relay, and so on. Once the sequence was
    > > complete, the list of values was compared for accuracy
    > > and the results displayed to an operator who passed or
    > > failed the power supply.
    >
    > > Regards,
    > > Dr. Diode
    >
    >
    > To UNSUBSCRIBE, just send mail to:
    > basicstamps-unsubscribe@yahoogroups.com
    > from the same email address that you subscribed. Text in the Subject and
    Body of the message will be ignored.
    >
    >
    > Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/
    >
    >
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-11 10:15
    Hi, can avoid using relays? Because it is the device that I am
    avoiding.

    --- In basicstamps@yahoogroups.com, PH <drdiode2002@y...> wrote:
    > Hello,
    >
    > I built a similar piece of test equipment a few years
    > back. It was an automated power supply tester, but was
    > very complex and used a PC and VisualBasic instead of
    > a Stamp.
    >
    > I used a GPIB equipped DMM for the measuring portion
    > of the project. I then built a multiplexor system with
    > high-quality relays to direct the desired measurement
    > at the appropriate time to the DMM. The relays were
    > driven by a digital I/O PCI card, and the GPIB
    > interface was a PCI card too.
    >
    > The program would increment the relay, hold it, take a
    > measurement with the DMM, store the value, increment
    > to the next relay, and so on. Once the sequence was
    > complete, the list of values was compared for accuracy
    > and the results displayed to an operator who passed or
    > failed the power supply.
    >
    > My program was designed to be quick and very automated
    > so as not to require a skilled operator. The operator
    > installed ten power supplies in bays one through ten,
    > and pushed start. Three minutes later, the operator
    > was instructed to pass or fail certain slots and
    > continue.
    >
    > If you are looking for something relatively slow, you
    > may be able to skip the GPIB and PCI section and
    > simply have a Stamp increment a relay matrix,
    > connected to a DMM, each time a button is pressed. The
    > operator (or yourself) would push the button, read the
    > DMM (one with low resistance capabilities, or course),
    > and press the button again, read it again, and so on.
    >
    > Cool project.
    >
    > Regards,
    > Dr. Diode
    > --- catcat <semiconductor@p...> wrote:
    > > Hi, I would like to make an automated test equipment
    > > to be used for
    > > measuring resistance of kelvin contactors.
    > >
    > > there ar 15 contactors so I want to measure the
    > > resistance of each
    > > contactors which is less than 1 Ohm. I will conduct
    > > a pin to pin
    > > measurement and a pin to ground measurement to check
    > > if the pins are
    > > grounded.
    > >
    > > Any ideas if how i can make this possible? The
    > > resistance involved is
    > > less than 1 Ohm.
    > >
    > > Thanks you.
    > >
    > >
    > > To UNSUBSCRIBE, just send mail to:
    > > basicstamps-unsubscribe@yahoogroups.com
    > > from the same email address that you subscribed.
    > > Text in the Subject and Body of the message will be
    > > ignored.
    > >
    > >
    > > Your use of Yahoo! Groups is subject to
    > > http://docs.yahoo.com/info/terms/
    > >
    > >
    >
    >
    > __________________________________
    > Do you Yahoo!?
    > Yahoo! SiteBuilder - Free, easy-to-use web site design software
    > http://sitebuilder.yahoo.com
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-11 10:16
    Hi, we are under the same problem, how todo the project w/oi using
    relays.

    --- In basicstamps@yahoogroups.com, "Richard Furniss" <rfurniss@e...>
    wrote:
    > Your right this is a cool project, the only problem I see is that
    you are
    > using a relay to measure relay contact resistance of another relay,
    even
    > though the selecting relay is high quality it doesn't sound good.
    >
    > Is there a non mechanical solution anyone can see.
    >
    > I'm working on a similar project and I worry about relay noise
    (clicking)
    > and relay life if you wanted to do continues monitoring.
    >
    > <snips>
    > >
    > > The program would increment the relay, hold it, take a
    > > measurement with the DMM, store the value, increment
    > > to the next relay, and so on. Once the sequence was
    > > complete, the list of values was compared for accuracy
    > > and the results displayed to an operator who passed or
    > > failed the power supply.
    >
    > > Regards,
    > > Dr. Diode
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-11 10:19
    Hi, can you gice a schematic of the circuit. Thanks.

    --- In basicstamps@yahoogroups.com, "Mike Sokol - ModernRecording"
    <mikes@m...> wrote:
    > For very low-resistance measurement, a good way to go is a 4-wire
    Kelvin
    > Bridge, which nulls out the resistance of the test leads
    themselves. That
    > way you can accurately measure resistance to below 0.1 ohms without
    having
    > to resort to jumper cables for test leads. It does this by
    comparing the
    > test resistance to a switchable bank of know-value resistors, and
    when the
    > needle in the meter nulls, the test resistance equals the comparison
    > resistance. We used this method to test and select resistors down
    to a few
    > ppm (parts per million) accuracy. If you need to switch banks of
    test
    > subjects, just use standard 4-pole relays.
    >
    > Mike Sokol
    > www.modernrecording.com
    > mikes@m...
    >
    >
    >
    Original Message
    > From: "Richard Furniss" <rfurniss@e...>
    > To: <basicstamps@yahoogroups.com>
    > Sent: Sunday, August 10, 2003 12:56 PM
    > Subject: Re: [noparse][[/noparse]basicstamps] Multiplexed resistance meter
    >
    >
    > > Your right this is a cool project, the only problem I see is that
    you are
    > > using a relay to measure relay contact resistance of another
    relay, even
    > > though the selecting relay is high quality it doesn't sound good.
    > >
    > > Is there a non mechanical solution anyone can see.
    > >
    > > I'm working on a similar project and I worry about relay noise
    (clicking)
    > > and relay life if you wanted to do continues monitoring.
    > >
    > > <snips>
    > > >
    > > > The program would increment the relay, hold it, take a
    > > > measurement with the DMM, store the value, increment
    > > > to the next relay, and so on. Once the sequence was
    > > > complete, the list of values was compared for accuracy
    > > > and the results displayed to an operator who passed or
    > > > failed the power supply.
    > >
    > > > Regards,
    > > > Dr. Diode
    > >
    > >
    > > To UNSUBSCRIBE, just send mail to:
    > > basicstamps-unsubscribe@yahoogroups.com
    > > from the same email address that you subscribed. Text in the
    Subject and
    > Body of the message will be ignored.
    > >
    > >
    > > Your use of Yahoo! Groups is subject to
    http://docs.yahoo.com/info/terms/
    > >
    > >
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-11 13:56
    There's a pretty good explanation of how a Kelvin Bridge works at
    http://newton.ex.ac.uk/teaching/CDHW/Sensors/#Resistance. Typically, a
    Kelvin bridge is used to compare a group of test resistors against a
    "standard resistor" to within a great degree of accuracy. This accuracy can
    be gained up or down at the sensing amplifier which drives the null meter.
    This "standard" resistor is normally calibrated against a known resistance
    value that's traceable to the National Bureau of Standards. Since you're
    only worried about nulling, the exact voltage used in the circuit doesn't
    really matter, and the exact sensitivity of the null meter doesn't matter
    either. Once you get zero current, the two resistor values must be equal.
    And you can add really long sense leads or insert things like relay
    contacts, and the circuit will still compare resistances properly. The
    disadvantage of a standard ohm meter type circuit is that it can be
    uncalibrated by any serial lead resistance, as well as the exact voltage
    applied across the test resistor, all of which is difficult to calibrate and
    verify to Government standards.

    Mike Sokol
    www.modernrecording.com
    mikes@m...


    Original Message
    From: "catcat" <semiconductor@p...>
    To: <basicstamps@yahoogroups.com>
    Sent: Monday, August 11, 2003 5:19 AM
    Subject: [noparse][[/noparse]basicstamps] Re: Multiplexed resistance meter


    > Hi, can you gice a schematic of the circuit. Thanks.
    >
    > --- In basicstamps@yahoogroups.com, "Mike Sokol - ModernRecording"
    > <mikes@m...> wrote:
    > > For very low-resistance measurement, a good way to go is a 4-wire
    > Kelvin
    > > Bridge, which nulls out the resistance of the test leads
    > themselves. That
    > > way you can accurately measure resistance to below 0.1 ohms without
    > having
    > > to resort to jumper cables for test leads. It does this by
    > comparing the
    > > test resistance to a switchable bank of know-value resistors, and
    > when the
    > > needle in the meter nulls, the test resistance equals the comparison
    > > resistance. We used this method to test and select resistors down
    > to a few
    > > ppm (parts per million) accuracy. If you need to switch banks of
    > test
    > > subjects, just use standard 4-pole relays.
    > >
    > > Mike Sokol
    > > www.modernrecording.com
    > > mikes@m...
    > >
    > >
    > >
    Original Message
    > > From: "Richard Furniss" <rfurniss@e...>
    > > To: <basicstamps@yahoogroups.com>
    > > Sent: Sunday, August 10, 2003 12:56 PM
    > > Subject: Re: [noparse][[/noparse]basicstamps] Multiplexed resistance meter
    > >
    > >
    > > > Your right this is a cool project, the only problem I see is that
    > you are
    > > > using a relay to measure relay contact resistance of another
    > relay, even
    > > > though the selecting relay is high quality it doesn't sound good.
    > > >
    > > > Is there a non mechanical solution anyone can see.
    > > >
    > > > I'm working on a similar project and I worry about relay noise
    > (clicking)
    > > > and relay life if you wanted to do continues monitoring.
    > > >
    > > > <snips>
    > > > >
    > > > > The program would increment the relay, hold it, take a
    > > > > measurement with the DMM, store the value, increment
    > > > > to the next relay, and so on. Once the sequence was
    > > > > complete, the list of values was compared for accuracy
    > > > > and the results displayed to an operator who passed or
    > > > > failed the power supply.
    > > >
    > > > > Regards,
    > > > > Dr. Diode
    > > >
    > > >
    > > > To UNSUBSCRIBE, just send mail to:
    > > > basicstamps-unsubscribe@yahoogroups.com
    > > > from the same email address that you subscribed. Text in the
    > Subject and
    > > Body of the message will be ignored.
    > > >
    > > >
    > > > Your use of Yahoo! Groups is subject to
    > http://docs.yahoo.com/info/terms/
    > > >
    > > >
    >
    >
    > To UNSUBSCRIBE, just send mail to:
    > basicstamps-unsubscribe@yahoogroups.com
    > from the same email address that you subscribed. Text in the Subject and
    Body of the message will be ignored.
    >
    >
    > Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/
    >
    >
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-12 18:13
    Hi Cat Cat:

    Many Digital Multimeters, like the Hp/Agilent 34401, have a 4 wire
    connection for resistance in addition to the two wire connection. There
    is a special set of leads for use with the Kelvin connection. The 34401
    reads out to 0.000,1 Ohms. The 4 wire, Kelvin, connection is very
    simple. One of the pairs has a current source and the other pair is a
    voltmeter. Suppose that all 4 wires had 10 Ohms of resistance and you
    want to measure a 1 Ohm resistor using a 1 ma current source.

    If you used a 2 wire Ohm meter then 1 ma * 1 Ohm is 1 mv, but the leads
    contribute 2 * 10 Ohms * 1 ma = 20 mv of error. You could try to use
    the delta DMM function by shorting the leads and pressing zero, but the
    lead resistance may vary as the leads are moved making it impossible to
    measure a resistance that's much smaller than the lead resistance. The
    contact resistance may also be greater or equal to the resistance of the
    device you're trying to measure.

    For a 4 wire, Kelvin, connection: The resistance in the voltmeter leads
    has an insignificant effect since the Voltmeter input impedance is 10
    Meg Ohms. Since the meter is connected on the device side of the lead
    resistance in the current pair that voltage drop is not seen by the
    voltmeter. The voltmeter ends up reading 1 mv. Note that if there is
    contact resistance and the voltmeter is connected closer to the DUT than
    the current pair the contact resistance will also be removed from the
    measurement. Also note that it really doesn't matter what the lead +
    contact resistance is, it will be removed from the measurement as long
    as the current source has the compliance to keep supplying the same
    current even though the voltage gets large.

    By measuring an accurately known resistor using the Kelvin connection
    you can determine the scale factor (calibration of both the current
    source and the voltmeter). This could then be used to improve the
    overall accuracy of the system.

    You can make your own current source from one of the National
    Semiconductor 3 lead temperatures sensors, a diode, as a few other
    parts. Use that with an A/D converter on the BS and you have a Kelvin
    measuring system. You could use DPST relays, but there are solid state
    switches that could also be used and the on resistance spec is not
    critical in the case of a Kelvin connection.

    Have Fun,

    Brooke CLarke, N6GCE

    >
    Original Message
    >From: "catcat" <semiconductor@p...>
    >To: <basicstamps@yahoogroups.com>
    >Sent: Monday, August 11, 2003 5:19 AM
    >Subject: [noparse][[/noparse]basicstamps] Re: Multiplexed resistance meter
    >
    >
    >
    >
    >>Hi, can you gice a schematic of the circuit. Thanks.
    >>
    >
    >
  • ArchiverArchiver Posts: 46,084
    edited 2003-08-12 21:11
    Also, for REALLY accurate measurements, an A.C. source current is used which
    cancels out any battery effects which result when dissimilar metals make
    contact. You really can measure resistances down to a tiny fraction of an
    ohm with the proper equipment.

    And as a side note, on a previous job we used a Kelvin bridge setup to
    measure spinning resistors sealed inside tiny glass tubes filled with Argon
    gas, which we would Laser trim to within 100 ppm of a target value. It was
    pretty cool....

    Mike Sokol

    Original Message
    From: "Brooke Clarke" <brooke@p...>
    To: <basicstamps@yahoogroups.com>
    Sent: Tuesday, August 12, 2003 1:13 PM
    Subject: [noparse][[/noparse]basicstamps] Multiplexed resistance meter


    > Hi Cat Cat:
    >
    > Many Digital Multimeters, like the Hp/Agilent 34401, have a 4 wire
    > connection for resistance in addition to the two wire connection. There
    > is a special set of leads for use with the Kelvin connection. The 34401
    > reads out to 0.000,1 Ohms. The 4 wire, Kelvin, connection is very
    > simple. One of the pairs has a current source and the other pair is a
    > voltmeter. Suppose that all 4 wires had 10 Ohms of resistance and you
    > want to measure a 1 Ohm resistor using a 1 ma current source.
    >
    > If you used a 2 wire Ohm meter then 1 ma * 1 Ohm is 1 mv, but the leads
    > contribute 2 * 10 Ohms * 1 ma = 20 mv of error. You could try to use
    > the delta DMM function by shorting the leads and pressing zero, but the
    > lead resistance may vary as the leads are moved making it impossible to
    > measure a resistance that's much smaller than the lead resistance. The
    > contact resistance may also be greater or equal to the resistance of the
    > device you're trying to measure.
    >
    > For a 4 wire, Kelvin, connection: The resistance in the voltmeter leads
    > has an insignificant effect since the Voltmeter input impedance is 10
    > Meg Ohms. Since the meter is connected on the device side of the lead
    > resistance in the current pair that voltage drop is not seen by the
    > voltmeter. The voltmeter ends up reading 1 mv. Note that if there is
    > contact resistance and the voltmeter is connected closer to the DUT than
    > the current pair the contact resistance will also be removed from the
    > measurement. Also note that it really doesn't matter what the lead +
    > contact resistance is, it will be removed from the measurement as long
    > as the current source has the compliance to keep supplying the same
    > current even though the voltage gets large.
    >
    > By measuring an accurately known resistor using the Kelvin connection
    > you can determine the scale factor (calibration of both the current
    > source and the voltmeter). This could then be used to improve the
    > overall accuracy of the system.
    >
    > You can make your own current source from one of the National
    > Semiconductor 3 lead temperatures sensors, a diode, as a few other
    > parts. Use that with an A/D converter on the BS and you have a Kelvin
    > measuring system. You could use DPST relays, but there are solid state
    > switches that could also be used and the on resistance spec is not
    > critical in the case of a Kelvin connection.
    >
    > Have Fun,
    >
    > Brooke CLarke, N6GCE
    >
    > >
    Original Message
    > >From: "catcat" <semiconductor@p...>
    > >To: <basicstamps@yahoogroups.com>
    > >Sent: Monday, August 11, 2003 5:19 AM
    > >Subject: [noparse][[/noparse]basicstamps] Re: Multiplexed resistance meter
    > >
    > >
    > >
    > >
    > >>Hi, can you gice a schematic of the circuit. Thanks.
    > >>
    > >
    > >
    >
    >
    >
    > To UNSUBSCRIBE, just send mail to:
    > basicstamps-unsubscribe@yahoogroups.com
    > from the same email address that you subscribed. Text in the Subject and
    Body of the message will be ignored.
    >
    >
    > Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/
    >
    >
Sign In or Register to comment.