ArchiverArchiver Posts: 46,084
edited 2002-02-05 07:04 in General Discussion
The data sheet for the DS1620 Digital Thermometer states that a higher
resolution can be obtained by reading the slope (the number of counts
per degree C for that temperature) and reading the Count Remaining, the
actual temperature may be calculated by:

Temp = Temp_Read - .25 + ((CountPerC - CountRemain)/CountPerC)

I tend to think that the Count Remaining should not exceed the Count
Per degree C (Slope). But that is not my case. My slope is fairly
stable ?? at 75 @ 20C, but will vary up to about 125. The Count
Remaining, however is all over the place. Between 30 to 500. And there
seem to be no relationship of having a slope of 75 and Count Remaining
of 30 or 500. Needless to say, the higher resolution is unusable.

Has anyone tried using the DS1620 in this matter ? Can it be done ? Any
available BS2 code I can try ? Can anyone shed some light on anything I
might be missing ?

Bruce Snowden

Comments

  • ArchiverArchiver Posts: 46,084
    edited 2002-02-05 07:04
    Hi Bruce,

    Here is some information and code using the DS1620 in high resolution mode:

    http://www.emesystems.com/OWL2face.htm#DS1620-HR

    You are right, the count-remaining should always be less than
    count-per-degree. There are a couple of fine points that might be
    bugging you. For example, the chip has to be operated in single-shot
    mode to use the high resolution mode.

    -- regards,
    Thomas Tracy Allen Ph.D.
    electronically monitored ecosystems
    mailto:tracy@e...
    http://www.emesystems.com


    >The data sheet for the DS1620 Digital Thermometer states that a higher
    >resolution can be obtained by reading the slope (the number of counts
    >per degree C for that temperature) and reading the Count Remaining, the
    >actual temperature may be calculated by:
    >
    >Temp = Temp_Read - .25 + ((CountPerC - CountRemain)/CountPerC)
    >
    >I tend to think that the Count Remaining should not exceed the Count
    >Per degree C (Slope). But that is not my case. My slope is fairly
    >stable ?? at 75 @ 20C, but will vary up to about 125. The Count
    >Remaining, however is all over the place. Between 30 to 500. And there
    >seem to be no relationship of having a slope of 75 and Count Remaining
    >of 30 or 500. Needless to say, the higher resolution is unusable.
    >
    >Has anyone tried using the DS1620 in this matter ? Can it be done ? Any
    >available BS2 code I can try ? Can anyone shed some light on anything I
    >might be missing ?
    >
    >Bruce Snowden