Need ideas to fix my ECG module. — Parallax Forums
chelletengchelleteng Posts: 29
edited 2012-04-13 06:36 in Propeller 1
Need some help/ideas to solve this issue I'm currently facing with my ECG module:
The TL084 is supposed to work as an amplifier and filter, it keeps giving irregular readings (not sure if it's unfiltered noise, etc) and it wouldn't give a negative value for the ECG reading.
Despite changing the gain at R2, the reading still does not differ much.
Thank you.
1024 x 1141 - 108K
    edited 2012-04-09 20:17
    Where is your 3rd electrode? typically there is a ground return to the body that allows the instrumentation setup to reject any common mode. In your schematic, this signal would typically be buffered from the mid-point of R2
  • HShankoHShanko Posts: 402
    edited 2012-04-10 12:55
    What type of ECG sensor are you using? Vendor source if purchased?
    Looks to be nearly what I need for my 'heart pulse display' using Raymans 3.5" touch LCD and pcb. Thanks for any info you can supply.
  • william chanwilliam chan Posts: 1,326
    edited 2012-04-11 06:52
    Beau,
    If the leg ECG sensor is connected to the mid-point of R2, would the voltage reference R8 and R9 still be required?
    Should C1 be replaced with a resistor to provide direct DC coupling instead of AC coupling?
    edited 2012-04-11 08:47
    I would focus on Figure #14 of this documentation. It's a modification from conventional ECG in that it improves current loop problems created in the leads going to/from the patient... Ra in the schematic could be simplified by halving R1.
    Just curious, with your current setup, what is the level of your noise floor? This needs to be extremely low in order to detect micro-volt levels picked up by the electrodes.
    edited 2012-04-12 05:31
    chelleteng,
    Could you post a picture of your TV screen to show us what could be the problem?
  • chelletengchelleteng Posts: 29
    When both the ECG inputs are short circuited:
    When the ECG inputs are attached to both wrists:
    When the ECG signals are not attached to anything (noise, etc):
    The heartbeat captured from the ECG circuit is very irregular. The number on the top left corner is the ADC reading from the Sigma-Delta converter.
    Any ideas? Thank you..
    1024 x 768 - 44K
    1024 x 768 - 42K
    1024 x 768 - 42K
    1024 x 768 - 47K
    1024 x 768 - 42K
    1024 x 768 - 45K
    1024 x 768 - 40K
    1024 x 768 - 46K