I have two SHT1x connected to two different BS2 boards running the sample code and both sensors are reading as much as ten to eleven degrees higher than actual room temp. One SHT is connected directly to a Home Work Board and the other is eighteen inches away from all other electronics.
Right now the SHT is reading 78.1F and 25%H.
Actual room temp and humidity are 75F and 31%H.
What can I do to get a true reading?
Thanks, Dave.
Right now the SHT is reading 78.1F and 25%H.
Actual room temp and humidity are 75F and 31%H.
What can I do to get a true reading?
Thanks, Dave.
Comments
The datasheet states:
What I have done is place the sensors on the window sill of my bathroom for a day or two. The steam from a few showers rehydrates the polymer glob and then I let them sit in a warm/dry environment for a day or two (I chose the top of my Wii console that is inside my entertainment center and is always warmer then the room). After that, all 4 sensors were within 2 degrees of each other and within 5% humidity of each other. (and my SHT15 reference sensor)
Keep in mind also, that the SHT11 is very sensitive, so keep the two very close to each other while testing them. Maybe place a small cup over them to prevent ambient airflow from disturbing the readings.
One of the sht sensors that has been in a 75% humidity environment for a week and the temp is +1° in comparison to the other instruments.
I don't know if this is because of the high humidity or that it has been running for a week. I still think that the rh reading is still 5% higher than actual rh but that might get better over time. The other sht has been siting un powered for a day in the same 75% humidity environment and I will test it tomorrow.
I also got a new sht in the mail today so I tested it too. The new sht reading is 81.9°.
I don't need any instruments to tell me that this room is not in the 80s.
So what is going on?
Thanks.
I borrowed an instrument from the lab today to see if I was wrong or the SHT was wrong. It's a Hygrometer, Thermometer, and Dew Point measuring instrument that has been calibrated and is due for calibration on 8/22/13. So I'm pretty sure that I'm starting out on solid ground with this instrument. I have my Parallax SHT and the other probe that also uses a SHT as a sensor about .5" from each other.
The readings from the STAMP and SHT are 71.4tF and 79.4rh
The readings from the certified instrument are 68.39tF and 76.78rh
That makes a 3.1 degree F difference and a 2.7 percent rh difference.
That seams like a big difference to me, is that an acceptable difference for an instrument that is supposed to measure to the tenth of a degree?
Am I expecting more than its capable of? I'll just stop asking questions if I am expecting to much.
Thanks for any help. I'm not trying to make anyone angry, just trying not to get shot down by my coworkers when I propose an idea.