Sensirion Humidity Sensor Question
Sal Ammoniac
Posts: 213
I've added a Sensirion humidity sensor to my weather station and frequently see humidity readings over 100% in the early mornings. It often goes as high as 106%.
When this occurs do I need to turn on the sensor's built-in heater until the humidity reading drops below 100%? Does a reading >100% indicate condensation on the sensor that can affect future readings?
When this occurs do I need to turn on the sensor's built-in heater until the humidity reading drops below 100%? Does a reading >100% indicate condensation on the sensor that can affect future readings?
Comments
It would not hurt to turn on the heater for a short time· but even then it may still read over 100 % when the sensor cool back down
▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔
··Thanks for any··that you may have and all of your time finding them
·
·
·
·
Sam
The accuracy of the sensor falls off at high humidity, +/-5% at 100% for the SHT11 according to the attached graph from the Sensirion documentation.
Are you using the recommended linearity and temperature compensation formulas?
I don't think the internal heater is going to help. It may have something to do with how the sensor is situated. The best location is inside a weather shelter or radiation shield or in the shelter of a tree, where it is shielded from radiation from the cold sky or from the ground.
▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔
Tracy Allen
www.emesystems.com
1) Is your code optimized for teh voltage you are supplying the sensor?
▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔▔
Andrew Williams
WBA Consulting
WBA-TH1M Sensirion SHT11 Module
Special Olympics Polar Bear Plunge, Mar 20, 2010
Propeller-Based Reverse Geo-Cache Birthday Present Project
Yes, I am using the recommended linearity and temperature compensation formulas from the datasheet. I have the Sensirion sensor mounted in a small plastic box about 2" x 1" x 1". I've drilled a grid of 1/8" holes in the metal cover of the box to admit ambient air. The box containing the sensor is mounted with the holes facing down in a standard cotton region wooden instrument shelter.
Andrew:
Yes, I am using 3.5v coefficients from the datasheet (I'm running the sensor at 3.3v and the 3.5v coefficients are the closest to that value).