calculate ppm of co2 module sensor
Dear Sir
I have a CO2 module sensor(#27929)
I take the different voltages between P1 and P2 to display ppm
Following op-amp calculate:
Vtp1_tp2 = Vin*(R1+R2)/R1
R1 = 150k
R2=1M
Following chart datasheet MG811, I choose
10.000ppm <---> 0.264mV---->Vtp1_tp2=2.02V
400ppm <----> 0.323mV
>Vtp1_tp2=2.48V
It means we have two points to draw the linear scale ppm
y = -20869.562x + 52156.5112
y:ppm
x: voltages of Vtp1_tp2
when test with x=1 and x=3.35. I have the value is very large and very small
x=1 --->y = 31286.95ppm
x=3.35 --->y = -17756ppm
I think my calculating is wrong.
Can you give me an advice.
Thanks you so much
R1 = 150k
R2=1M
>Vtp1_tp2=2.48V
y:ppm
Thanks you so much
If your original calculations are correct (for Vtp1_tp2), then your formula may be correct. It may not be valid for voltages much more than 2.48V nor for voltages much below 2.02V. Look at the MG811 graphs. The output voltage of the sensor isn't going to be much below 0.250V nor above 0.350V.
Also, there is lots of variation from sensor to sensor. The good folks on the Arliss project (Stamps-in-class forum) last year compared quite a few of them on one graph. I can't find that at the moment. The upshot is that each sensor has to be individually calibrated, at least one point (atmospheric).
EMF = E0 - R*T*logPco2/(2*F)
Li2Co3----> 2Li+ + Co2 + 1/2O2 +2e- : E0 = 3.05
2Na+ + 1/2O2 + 2e- ---->Na2O :E0 = -2.71
LI2CO3 + 2Na+ ---->2Li+ + Na2O + CO2 :E0=3.05-1.71=0.34
R gas constant = 8.31(volt-colomb/mol)
F faraday =96500 colombs/mol
E0=0.34
PC02 ppm of CO2
EMF (volts)
T: absolute temperate(K)
the result of co2 ppm
PCO2 = 10exp(EMF*2F/(RT))
The question is
T is a surface temperature in MG811 or air temperature.
If T is temperature in MG811 how can we measure it.
thanks Tracy Allen. Arliss project help some helpful information.
Anyway, you can't use the formula directly. It will depend on calibration at a reference concentration. Say, ambient at ~390ppm or compared with a better instrument.
Find EMFref at that reference concentration, Pco2ref, then when reading at a different concentration, (EMFx, Pco2x):
EMFx - EMFref = M * (log(Pco2x) - log(Pco2ref)) = M * (log(Pco2x / Pco2ref))I rolled R*T and a fudge factor into the single constant M. If you do a single point calibration, assuming for example Pco2ref = 390 ppm in the atmosphere, then you can eyeball the slope of the graph in the MG811 data sheet (semilog graph of EMF against log ppm) and use that for M. Better, find a more concentrated CO2 source to use for a second calibration point or for a series of dilutions.